Mastering Gas Chromatography: Unlocking Efficiency with Sample Conditioning and System Design

Learn best practices for troubleshooting, minimizing time lags, and safeguarding your system against common pitfalls while ensuring safe, accurate sample analyses.
Oct. 16, 2024
14 min read

Understanding the Sample Port

The core objective of the sample port is to draw a representative sample from the process. A good design should avoid stagnant regions, flow disturbances and two-phase areas.
 
You should factor several items into the sample port design, including process temperature, pressure, stream velocity, size of the pipe and physical properties of the sample makeup, including corrosivity, toxicity and flammability or reactivity. Although you may not need a filter at the tip of the sample probe for clean streams, it’s a good precautionary practice. 
 
Other important factors to consider:
 
  • Don’t design a sample port on control valve bypass, drains or vents, level bridle and other dead spots. If the sample port needs to be installed on a large vessel or a reactor, make sure the area near the sample probe is well-mixed. 
  • Flow obstructions such as ells, reducers, trees, control valves, block valves, thermowells or orifice plates or pitot tubes cause flow disturbances, which could cause sample distortion. Typically, sample probes require 10 pipe diameters of straight, unobstructed pipe upstream and five diameters downstream, though exact requirements vary with flow characteristics and obstruction type. When extracting a sample from a vessel, tank or reactor, make sure the sample pot is in a well-agitated zone. 
  • Probes protruding in the pipe are preferable to probes that are flush with the pipe wall.  This is particularly important if the process stream is laminar with static film. In large pipes with high turbulent flows, ensure the sample probe has sufficient mechanical support. 
  • The probe opening should face the flow. Avoid sample probes entering from the bottom of a horizontal pipe or vertical pipe with downward flow. 
  • While stainless-steel probes are common in many applications, highly corrosive processes may require exotic metallurgies. Consider not only the main constituents of the sample, but low-level, highly corrosive impurities as well to ensure that the probe will not be affected by corrosion.  As a quick guide, probe metallurgy should match with that of the process vessels or piping. 
  • For pipeline sample ports, avoid bottom entry for a horizontal pipe. Avoid sample ports on a vertical line with downward flow. For vapor samples, if you anticipate significant liquid entrainment, you may need to consider deflectors or chevron separators to remove liquid. 
  • Provide for safe isolation of the process from the sample system. For processes at very high pressures, consider double block-and-bleed arrangement. Similarly, for extreme temperatures, consider appropriate insulation. 

Sample Conditioning’s Role in GC Optimization

The sample conditioning system within a GC prepares the material so it’s suitable for testing with the proper temperature and pressure. In addition, this system should minimize the need for components, such as coalescers, that could delay delivery of the sample to the GC. 
 
Membrane filters within the system help remove liquid and only allow vapor to enter the GC (1). Operating experience has shown that most GC system issues arise from the design, installation and maintenance of the sample-conditioning component of the GC system. 
 
It’s important to use appropriate pressure, temperature and delta-pressure transmitters to help with troubleshooting. In recent years, the new sampling/sensor initiative encourages development of miniature sampling systems with relevant sensors for flow, pressure and temperature and their interface with diagnostic capabilities of GC controller (2). 
 
During sample conditioning, consider these critical factors: 
  • Remove liquid droplets and or particulates if they are not a part of the analytical requirements. However, if droplets and/or particulates are a part of the analytical requirements, then liquid will need to be vaporized. Heavy molecular liquids and particulates are not suitable for GC analysis, and they will need to be analyzed by different techniques such as liquid chromatography. 
  • Fine filters, say 1 to 2 microns, can be used to remove particulates. In cases where the proportion of undesirable particulate entrainment increases, you will need consider strainers to remove some particulates before sending the sample to the fine micron filters. 
  • Most GCs operate at 20 psig to 30 psig. For high-pressure processes, use pressure regulators (reducers) with appropriate safety systems. Also, provide block valves with a tight shutoff to isolate the process from the sample system. 
  • If the process involves chemicals, such as light hydrocarbons, that could undergo Joule-Thompson cooling effect, ensure you have adequate heating or temperature-controlled heat tracing to prevent condensation. Effective steam trace requires no accumulation of the condensate in the line. Electric heating should make close contact with the sample line. 
  • A pressure booster/blower must be provided for processes at low pressure or vacuum. With vacuum systems, you need to consider the possibility of air ingress.
  • Electrical-area classification requirements may force the GC systems or their shelters to be installed in non-rated areas, which could be far from the process. This will, in turn, increase the time lag in reaching the GC. Fast-loop arrangements may be required in some situations. In a fast-loop arrangement, a slip stream from the process is routed closer to the GC, where a sample is withdrawn. This, in effect, reduces the time lag between the process and the analyzer.
  • Obviously, carrier gas supplied from gas cylinders must be maintained at a set pressure and flow rate with a specified purity level. Excessive moisture or impurities will affect GC reliability and expected useful life. When the gas cylinder needs to be switched, be aware of the need to purge air pockets, which if left in the sample line, could cause nuisance peaks.

Columns, Sample Valves, Controllers and Shelters

 
Column vendors provide guidelines on operating parameters of their columns. GC columns are typically coiled 316 stainless-steel tubing with varying length in a diameter of ⅛ in.-¼ in. 
 
The sample or switching valve is a rotary valve, which has typically six ports driven by solenoids controlled by the GC controller. This introduces a very small, typically microliter, volume of the sample along with an inert carrier gas. The controller also controls sequencing of valve operating modes including On/Off and makes it possible to isolate the GC column. Because the rotary valves have very small diameter ports and grooves, they are prone to plugging/clogging by fine dust particles. 
 
That’s why it’s critical to filter fine particles (1 to 2 micron) before sending the material to the sample valve. The sample, along with the carrier gas, is called moving phase. The material contained in the GC column is called stationary phase. There are two variations of the GC column: packed columns and capillary columns. Packed columns contain a solid packing coated with a layer of active substrate that absorbs chemicals from the sample. 
 
Capillary columns, although available in several versions, typically have the active substrate coated on the walls of the capillary or on a layer of inert film of solid support. Capillary columns have much less pressure drop than packed columns. 
 
Components with high solubility tend to stick to the substrate quickly, while those with lower solubility will flow further down the column before being absorbed (or adsorbed) on the substrate (Figure 2). The duration for which a particular chemical stays in the column is called retention time.
 
Consider the following:
  • GC column length depends on the granularity of analyses. As you increase the number of components for detection, column length increases.
  • Poor temperature control of the GC oven could cause a shift in retention time and result in erroneous analyses reported by the system.
  • With use and depending on your application, a GC column could accumulate heavy chemical species, which could result in ghost peaks.

Selecting a Detector

Sample components segregated by a GC along with the carrier gas are sent to detectors that use different physical or chemical properties (e.g., thermal conductivity, flame ionization, electron capture, mass spectrometry and others) to create a signal proportional to the concentration of specific components. 
 
You select the detector based on your sample constituents and their composition level and range (percent, ppm, ppb or other) accuracy of analysis, sensitivity, linearity and speed of response, interference on analysis by impurities or other chemicals, environmental regulatory requirements and maintenance considerations and collective experience of your plant personnel. 
 
Flame ionization (FID) and thermal conductivity (TC) detectors are used predominantly. Each detector has its plusses and minuses. Some broad considerations applicable to all detectors include:
 
  • Make sure all functional requirements for proper functioning of the detector are met. For instance, hydrogen/air used for FID must have the required purity, pressure and flow rate. Since thermal conductivity (T/C) of a gas is influenced by temperature, it is important to ensure that the T/C has set temperature. 
  • Periodic maintenance of the detector and ancillary equipment are vital for high reliability.
  • During the selection and operation of detectors, pay attention to potential interference by impurities. The concentration of chemical species of interest and linearity of response also determine the type of detector suitable for your application. For example, electron capture detectors would be preferable to FID for determining concentration of chemicals in the ppb (parts per billion) range. 
  • Provide for safe venting/disposal of gas stream after it comes out of the detector chamber. Gas venting must be in a safe location, away from congested areas. If the gas is vented to a flare header, provide a check valve on the sample exhaust to prevent contamination from backflow. 
  • Diagnostics, relevant coded messages and settings are often vendor specific and equipment manual should be consulted. 

Flame Ionization Detectors

These detectors are preferred for analysis of hydrocarbons, CO2 and CO. Note, CO and CO2 require reduction to methane before sending them to FID. 
 
The FID detector (Figure 3) contains an igniter or a filament that ignites the mixture of hydrogen air. Combustion generates ions in proportion to the concentration of the chemical. Ions contact collector plates or electrodes and generate electric current. 
 
Although FID by itself does not identify a specific chemical (it only measures current), the GC computer converts the signal to concentration of specific chemical specie based on the response factor, area of the peak (Figure 4) and the retention time in GC of that chemical. 
 
Salient points to keep in mind during design/ installation:
  • Make sure hydrogen and air with the required purity, ratio, pressure, temperature and flow are provided to the detector. Follow purity specifications from the vendor. Not meeting these requirements can cause the flame to be unstable or flame may not stay lit.
  • At the start up, if the temperature of the igniter (filament) is too low, it will not sustain flame. Typically, the temperature of the filament is kept 40–45 °F above the GC temperature to ensure stable flame during light off of hydrogen. With use, FID detectors would need to be cleaned periodically. 
  • Gas coming out of an FID should be disposed of at a safe location. In addition, hydrogen is a highly flammable gas; good ventilation and air monitoring is vitally important. Similarly, during maintenance/troubleshooting an FID, keep in mind the igniter may be hot to the touch and consider hydrogen flow a part of safety precautions. 

Thermal Conductivity Detectors

These detectors are based on the difference of the thermal conductivity of the carrier gas or reference gas and the gas carting the chemical constituent coming out of a GC. Reference gas flows across a resistance and depending on the thermal conductivity (heat-carrying ability) of the gas will determine the temperature of the resistor and the voltage. 
 
Although TC detectors are used for inorganic applications where FID is not applicable, TC is also used for hydrocarbons in the ppm to percent range. 
 
Key points to keep in mind include:
  • Make sure there is unimpeded and leak-free flow of gas from the GC, reference gas with the required purity, pressure and flow and electrical connections secure with proper polarity. Obviously, lack of adequate purity could affect baseline, analysis and could even damage the resistor (filament). 
  • Although helium (He) is the preferred carrier gas, supply shortages may require you to change to another gas, say hydrogen or nitrogen. Because of significantly higher thermal conductivity of He relative to other chemical compounds, it gives excellent sensitivity in detecting chemical constituents. 
  • Corrosive or fouling species will tend to destroy resistors. Baking the TC should help remove the fouling deposits. If that fails, the resistors will need to be replaced. 

About the Author

GC Shah

GC SHAH, PE, CFSE, CSP, CFPS, is a Houston-based consultant specializing in process safety, including hazard analysis and fire protection services. 